Hi, I have calibrated an US probe with "UF" image orientation. Thus, I have an I would like to pick points on the image and apply ImageToProbe transform to have their coordinates in the probe reference. As I see in http://perk-software.cs.queensu.ca/plus/doc/nightly/user/CommonCoordinateSystems.html, the image coordinates represents the pixel position in an "MF" oriented image. Does it mean I have to some extra flip on my picked point before applying Thanks a lot, |
Beta Was this translation helpful? Give feedback.
Replies: 1 comment 3 replies
-
Hi, the image orientation in the device definition is there to make the image coordinate system consistent with the hardware, the physical ultrasound probe. If you check the output of PLUS and the images are not in MF (the coordinates of the first axis do not increase towards the physical mark on the ultrasound probe, etc.) then you can still track your ultrasound correctly. The calibration will compensate for the wrong image orientation definition, and you won't experience any practical errors. But on the long run, it is recommended that you set up image orientation in the device definition correctly. That way your Image coordinate system will be consistent with all other PLUS-based systems. |
Beta Was this translation helpful? Give feedback.
Hi, the image orientation in the device definition is there to make the image coordinate system consistent with the hardware, the physical ultrasound probe. If you check the output of PLUS and the images are not in MF (the coordinates of the first axis do not increase towards the physical mark on the ultrasound probe, etc.) then you can still track your ultrasound correctly. The calibration will compensate for the wrong image orientation definition, and you won't experience any practical errors. But on the long run, it is recommended that you set up image orientation in the device definition correctly. That way your Image coordinate system will be consistent with all other PLUS-based syst…